#1 Posted : 09 August 2017 03:37:38(UTC)

Joined: 11/05/2017(UTC)
Posts: 1

Hi I would like to know what is the largest column can be used in the AKTA pure 150 system? Can I pack and run AxiChrom column in this system? Best regards!
GE Protein purification team  
#2 Posted : 09 August 2017 11:18:40(UTC)
GE'GE Protein purification team

Joined: 16/11/2015(UTC)
Posts: 106

Thank you for Asking. ÄKTA pure 150 supports columns up to 50 mm diameter including the AxiChrom 50 columns. Depending on the media, 70 mm columns may be used if linear flow rate and pressure allows for this. However, ÄKTA pure does not support packing of AxiChrom columns. ÄKTA avant 150 includes pre-made methods for intelligent packing of AxiChrom columns up to 70 mm diameter. The intelligent packing support includes pre-defined settings adapted to different media and ensures a controlled packing of the column. Hope this clarifies. Lotta, Hedkvist, Global Product Manager, Lab scale ÄKTA systems